air bubble appear in the sample syringe during purge and there is vaiation in peake area.how to solv

Options
<p><span style="font-size: 12pt;"><em>air bubble appear in the sample syringe during purge and there is vaiation in peake area.how to solve this problem</em></span><span style="font-size: 12pt;"><strong><em>?</em> </strong></span></p>

Best Answer

  • rune
    Answer ✓
    Options

    Hi

    I think it will be easiest if you remove the syringe and fill it with methanol and the reinstall it. Please se the attached procedure on how to do this.

    Best regards

    Rune

Answers

  • rune
    Options

    Hi

    While running 5-10 purges of the sample syringe lightly tab the syring to get the air buble to move to the top so it can be flushed out. If this does not solve the problem loosen the syring while purging.

    As a last resort remove the syringe, fill it with methanol , remove any air and attach it again and run the purge.

    best regards

    Rune

  • mmarie
    Options

    Many thanks Rune,

    but can you please explain how to loosen the syringe.