he signal-to-noise ratio
Options
According to Usp in acetaminophen methodology, he asks me for a 4-Aminophenol sensitivity test where I must show no less than 10 for 4-aminophenol in the signal-to-noise ratio. Someone can help me how to get this result in empower 3 and report it.
0
Best Answer
-
You will need Suitability tab in processing method and 'Calculate USP,EPand JPs/n ticked. Leave others as default. Click on alter sample table and go to table properties to show blank column. Tick blank (must acquired before sensitivity injection) then save and process . S/n should be now calculated. Report in table properties click 'sys. suti . Peak and choose SS as required0
Answers
-
This article in the Waters Knowledge base might also help - How to calculate USP and EP Signal to Noise in Empower.
Good Luck!
Marc1