Dear friends,
Today i observed "Restriction missing" when running the purge injector on Alliance 2695 HPLC system. I connected column and with the column i m getting 2000 pressure with 1.5ml/min flow.I done Several times purge injector by enabling compression check. problem is still persists.
could you please give some guidance in order to resolve it?
Today i observed "Restriction missing" when running the purge injector on Alliance 2695 HPLC system. I connected column and with the column i m getting 2000 pressure with 1.5ml/min flow.I done Several times purge injector by enabling compression check. problem is still persists.
could you please give some guidance in order to resolve it?
0
Answers
-
Compression checks require a column or a back pressure regulator. You need more than 2000psi, so get a similar column with smaller packed particles or add another column.0
-
0
-
0
-
Dan - Thanks for your response but I'm afraid it is not quite accurate. The compression check requires a back pressure of at least 500 psi but there is some evidence that pressures above 2000 psi can cause an issue. There are a couple of articles in the Waters Knowledge Base that explain and provide some potential solutions. This link should show you some of those - https://support.waters.com/Special:Search?fpid=230&q=%2Brestriction+%2Bmissing&path=&limit=55&page=2&tags=
Good luck Ravihans!0