Im having a strange problem with Empower 2. I set a value in the Limits tab for a peak custom field called Adjusted_Value (which is Amount/CConst1*100) and I set a Target of 100 with a lower warning value of 50. I have Ignore Blank values set to yes and check limits set to Always.

Weirdly , when I processed a run I got a lot of values which were way lower than 50 and it came up red and italic in review and a peak table in report but yet the field Faults was No in the report! I work with Faults in Empower 3 and anything which is outside of the limits comes up as Yes for the Faults field..I tried a different processing method, thinking it was corrupt but same thing- values which were way lower than the lower limit were faulting font correctly but yet the Faults field is still saying No!


Does anyone know what else could be causing this? I'm sure it cant be an Empower 2 anomaly as it would be well known. I'm not having any issues with Fault appearing as Yes in Empower 3. its probably something small but I cant think of anything...

Best Answer

  • HeatherLongden
    Options
    Lower Warning levels are only used to put lines on a control chart, they wont create faults. You have to put the limit in the ERROR level (or some field other than the Warning Level) to get faulted data. I'm surprised however, that it showed as red like a fault when using a warning level.

Answers

  • jkelley
    Options
    Hi,

    Have you looked for answers to your question in our Waters Knowledge Base? Lots of great information in there. You can even find KB articles by searching for them using a search engine like Google.

    Check out support.waters.com

    Regards,

    Justin Kelley
  • Ah thanks Heather, I thought both error and warning would fault values and was wracking my brains thinking what I had done wrong!
  • It’s an enhancement request for sure. Specifically to help with OOT monitoring.