MAX of 3 different Custom Fields
Options
Hello, it is possible to Calculate the Maximum of 3 different CFs?
For example:
CF1 = Result, Real(0.0)
CF2 = Result, Real(0.0)
CF3 = Result, Real(0.0)
MAX = MAX(CF1,CF2,CF3) ???
thank you
0
Answers
-
No, max is a summary function and can not be used in this manner.
Summary functions must have the formal X.Y.Z.FUNCTION(Parameter), where
X=sample label
Y=sample injection
Z=detector channel
FUNCTION = AVE, Min, MAX, etc, etc.
Parameter = detector channels, peak height, peak area, etc, etc.
0 -
You could do this in a report method if that suits. Have your 3 results in a table and add the MAX function to this table.
0 -
If you are set on a custom field to do the comparison and you wish the result to be whatever the max is, possibly an enumerated list with boolean fields could work?
Something like:
ENUM(GT(CF1, CF2)>(CF1, CF3), GT(CF2, CF1)>(CF2, CF3), GT(CF3, CF1)>(CF3, CF2))
translation table set to field with something like the following in the translation table for where CF1 is greatest: CF1 (fc) and so on for the others.
1