How to calculate amount without quantifying
Options
If i weigh out 10mg into 100mls of a 100% pure standard and put the concentration into the Value column which would in this case be 0.1mg per ml then inject my standard how do i find out exactly what the amount is without using bracketed standards? I presume the Calc Value is what you actually recover? Is the Cal Value a field that can be added to a report?
0
Best Answer
-
Yes, the calculated value is generally shown as a part of a calibration table. If you look at the report editor, under the group LC Calibration Point Tables, there is an item called Point Information. This will give you a standard table showing the calculated values. Also included in this table is a recovery of sorts, by means of a % deviation column.
0
Answers
-
Thanks for that shaunwat, I couldn't see it anywhere in the report.0
-
You'll need to look in the calibration curve fields to find it I think,0