BEH C18 Column high pressure issue
Comments
-
Hi Seamus,
I'm not clear if you saw this pressure rise without doing a sample injection or if you injected your sample and then ran the gradient. If you have injected a sample then I would suspect that it is your sample that is causing the clogging. If not then something in your mobile phase (as you have said) or something in the plumbing for the A line is causing the problem. Do you add anything to your water? Have you changed the solvent filters? Did the engineer replace seals etc. or not? These are things I would check but you may already have done them. Good luck. Please report back.
0 -
I believe it was the filters i was using. Today i didnt filter and removed the inlet frit in the column inlet and had a perfect run there after...fingers crossed it is!
0 -
You can try to install an inline particle filter between the injector and the column. I have installed one on my UPLC. I also put an inline particle filter in the fitting before every guard cartridge and column. It is much cheaper to replace the particle filter. Besides Waters inline guard cartridge, you can look into Optimize Technologies, or Analytical Sales & Products Inc for more user-friendly equivalent .
You may still want to protect the pump with solvent titanium filter, and consider replacing the solvent filters more frequently or clean them by sonication.
Good luck.
Albert