H-Class FTN Sample manager - How to calibrate Z-Axis?
<p> I can't find written instructions anywhere on how to do this. No help files, no PDFs on-line.</p><p></p><p>Help a brotha out!</p>
0
Answers
-
Nevermind.
It was in the help file after all. I missed it the first time.
0 -
I see you found the information. For others benefit this can be found in the ACQUITY Console under the Sample Manager FTN Maintenance pull down menu the choose Calibrate Z Axis. You will need to remove the well plate and using the Z buttons slowly bring the needle down very close to touching the tray bottom, then press Save.
Regards,
0 -
Same as with the regular Acquity, thanks