H-Class FTN Sample manager - How to calibrate Z-Axis?

<p> I can't find written instructions anywhere on how to do this.  No help files, no PDFs on-line.</p><p></p><p>Help a brotha out!</p>

Answers

  • Nevermind.

    It was in the help file after all.  I missed it the first time.

  • I see you found the information. For others benefit this can be found in the ACQUITY Console under the Sample Manager FTN Maintenance pull down menu the choose Calibrate Z Axis. You will need to remove the well plate and using the Z buttons slowly bring the needle down very close to touching the tray bottom, then press Save.

    Regards,

  • Same as with the regular Acquity, thanks